Ok the IR unit is getting power, it is just not warming up. Can that go bad. Seems like it's a solid state piece of equipment. It is getting 120 ac v across it
Ok so HC shows no hertz reading at all BUT can adjust it when the zero button is pushed in its rock steady. But when I do the span button nothing will not move.
Now the CO gives a reading of about 34 hertz. BUT can not adjust nothing, zero or span. Ohmn checked the leads and all are good.
It's only the CO that is not able to be adjusted. And as you can see only the two paper tags are still on the unit. Also this was inspected Nov 6th 1975 and Aug 11th 1979. So this unit is minimum 50 yrs old
No casting marks can be found.
Looked for a patent number, no luck.
Unless it does not say "patent no#"
There are a few numbers on paper stickers but nothing that looks like a patent number.
Well I did manage to find the pn
Now I just need to find the parts.
I have an operating manual which has a very simplified wiring schematic in it.
Pn 7049-0961
Thanks that was the ticket. But now that all is hooked back up, I now have a noisy CO gauge. Can not get it to zero out at all. I guess I have a bad IR chip. Now have to run one or two of those down