AC clamp Meter. What's it actually reading?

seagull369

Well-known member
Joined
Jan 16, 2013
Messages
227
This question is in regards to a true-RMS clamp meter set on amps ac,,, On the non-inrush setting, is the value displayed considered to be a true-RMS reading of the current waveform it's analyzing? When on the inrush current reading, is that simply just giving you the largest peak current (I-max) value it's able to capture?
 
Last edited:
To avoid these ads, REGISTER NOW!

u2slow

Well-known member
Joined
Nov 20, 2011
Messages
3,585
Location
BC
Your first question.... A true-rms meter gives you an true-rms reading. That's the whole point.

The second.... you have to find out how the manufacturer defines the feature. I suspect it will be the RMS value of whatever inrush it captures.
 
To avoid these ads, REGISTER NOW!

RPH

Well-known member
Joined
Dec 17, 2006
Messages
4,190
Location
Michigan Thumb
As with anything electrical math is involved. Most use a resistor used as bourdon resistor across transformer windings. This measures a voltage drop across it. The math is required for the types of non-sinusoidal waveforms.

DB32F69C-B641-4044-A6A2-A0749660CAD5.gif

9C46FD7B-1DFC-4E69-8C35-7B71128A5E0A.jpeg
 

dave*99

Well-known member
Joined
May 5, 2009
Messages
4,259
Location
Coastal NJ
As with anything electrical math is involved. Most use a resistor used as bourdon resistor across transformer windings. This measures a voltage drop across it. The math is required for the types of non-sinusoidal waveforms.
I think you mean burden resistor.
Bourdon invented the Bourdon tube for measuring gas pressure.
 

dave*99

Well-known member
Joined
May 5, 2009
Messages
4,259
Location
Coastal NJ
As with anything electrical math is involved. Most use a resistor used as bourdon resistor across transformer windings. This measures a voltage drop across it. The math is required for the types of non-sinusoidal waveforms.

DB32F69C-B641-4044-A6A2-A0749660CAD5.gif

9C46FD7B-1DFC-4E69-8C35-7B71128A5E0A.jpeg
Thankfully the true RMS meter does the math. Often the waveform that must be measured is a combination of more than one of those neatly drawn waveforms.
The true RMS meter digitizes the signal and calculates the reading.

A simple way to think of RMS is it is the heating value of the waveform.

If you connected a resistor to a DC source and measured its temperature then switched to a complicated AC waveform and the resistor rose to the same temperature…… you would know the RMS value was the same for both.
 

rlitman

Well-known member
Joined
Oct 18, 2010
Messages
24,591
Location
Long Island
Is your meter an AC only clamp meter, or an AC/DC current clamp. They work fundamentally differently.
Also, most meters sample more quickly than they refresh their displays, so your inrush measurement will likely be higher than a max-hold type current reading, if your meter has both modes.
 