The unit for the counts depend on the resolution of the range you are in. Whatever range, it's 20 counts on the end. For example, say you are measuring 1.014 volts, then your counts would be +/- 0.020V. Now if you go up a range and read 01.01V, your counts are now +/-00.20V a bigger change. I'm curious which meter you saw has an error of +/- 20 counts. Usually, that's the sign of either a bad cheap meter or a meter that is really high resolution. For example, my UT61E has a resolution of 22,000 counts and the worst counts is 10 on AC voltage. Similarly, my Brymen BM869S has a 500,000 count DCV range where they are really pushing the Analpg-Digital Converter to its limits and on that one you can see 20 or 30 count errors. That being said, the resolution is so high that there error is still very small. On the other hand, the UT210e has a much more reasonable single digit counts, as expected with it's resolution.Are the counts in mA as well?
Looking at the accuracy of the clamp meter:
Clamp meters calls out:
Range: 0-5,000mA
Resolution: 1mA
Accuracy: +/- 2.8% +20 digits
If I had a reading of 32.6 then my accuracy would be (32.6mA*0.028)
Therefore my reading would be 32.6 +/-0.9128mA +/- 20 counts. To get the total accuracy do I add the counts does that make it 32.6 +/- 20.9128mA. That just seems like a big difference in numbers. My reading of 32.6 could be anywhere between 11.7mA and 53.5mA. That could be the difference of a car having a parasitic draw vs. not having one.
I just wanted to double check to make sure the math is correct.