I have the same manual problem is the three way valve does not point out which port goes to what. The low pressure port which port do you use top or bottom
Oh, sorry. I don't have one anymore but will look around and see if I can find any. I did take a bunch of pics at one point. Did you search on here? I know there were a couple threads about them.I have the same manual problem is the three way valve does not point out which port goes to what. The low pressure port which port do you use top or bottom
That's why I asked for pics.
Yes please take pics of all the vacuum lines top and bottom. The back side between the bowls are still there."Tomorrow" meant, later today I should be able opening such unit, taking pictures of whatever required, inside that unit.
Please specify what you would like to see...
Thanks, I will lookOh, sorry. I don't have one anymore but will look around and see if I can find any. I did take a bunch of pics at one point. Did you search on here? I know there were a couple threads about them.





Thanks that was the ticket. But now that all is hooked back up, I now have a noisy CO gauge. Can not get it to zero out at all. I guess I have a bad IR chip. Now have to run one or two of those down
You should be able to put an oscilloscope on the detector signal somewhere on the board. That's how it's done on an Allen machine to determine if it's a detector fault.Thanks that was the ticket. But now that all is hooked back up, I now have a noisy CO gauge. Can not get it to zero out at all. I guess I have a bad IR chip. Now have to run one or two of those down
If I only had a scopeYou should be able to put an oscilloscope on the detector signal somewhere on the board. That's how it's done on an Allen machine to determine if it's a detector fault.
Looked for a patent number, no luck.By the way: there will be a patent number to be found on the optical unit. Google that and find plenty information on how this thing works....
By the way: there will be a patent number to be found on the optical unit. Google that and find plenty information on how this thing works....


The frequency function on a multimeter might give some indication. If you get a signal then you know the IR source is working and the detector is at least partially working.If I only had a scope
Ok will try thanksThe frequency function on a multimeter might give some indication. If you get a signal then you know the IR source is working and the detector is at least partially working.
And this is what? Patent number?
Ok so HC shows no hertz reading at all BUT can adjust it when the zero button is pushed in its rock steady. But when I do the span button nothing will not move.Ok will try thanks
Check the function of the switches and check that the IR sources get power.Ok so HC shows no hertz reading at all BUT can adjust it when the zero button is pushed in its rock steady. But when I do the span button nothing will not move.
Now the CO gives a reading of about 34 hertz. BUT can not adjust nothing, zero or span. Ohmn checked the leads and all are good.
Within the optical unit, between the tubes, there is a white phenolic block. On top it shows a screw and a small tube or some round metal thing.check that the IR sources get power.
Ok the IR unit is getting power, it is just not warming up. Can that go bad. Seems like it's a solid state piece of equipment. It is getting 120 ac v across itCheck the function of the switches and check that the IR sources get power.
The switches are known to play up on Sun machines.
I don't believe the IR source heats up, the heating up is a different part but I'm not even sure if the EPA-75 would have a heater.Ok the IR unit is getting power, it is just not warming up. Can that go bad. Seems like it's a solid state piece of equipment. It is getting 120 ac v across it
Its a cartridge heater, it glows red hot, they do burn out with age you can get a replacement from the watlow company The I.R. bench was made by Infrared industries for Sun, Allen, Bear The HC. and CO detectors tend to get noisy...meter needles jumping, with age. I think temp and humidity causes them to degrade. They are a I.R. sensitive photo cell Lead/Selenide chip, with a glass filter over themOk the IR unit is getting power, it is just not warming up. Can that go bad. Seems like it's a solid state piece of equipment. It is getting 120 ac v across it